  1 \n    Sound demodulation is active.    \n    Turn Sound demodulation off.     
  2 \n  Vertical scale is set to Linear.   \n      Set the scale to dB/div.       
  3 \n      Preselector is turned on.      \n        Select manual tuning.        
  4 \n      Cannot change parameter.       \n  Wait until Preselector is tuned.   
  5 \n        Span is set to 0 Hz.         \n          Change the span.           
  7 \n     Trace mode is set to Blank.     \n        Change to Write mode.        
  8 \n  Not available for baseband freq.   \n    Move marker before executing.    
  9 \n    Power measurement is active.     \n     Turn power measurement off.     
 10 \n       Signal track is active.       \n       Turn Signal Track off.        
 11 \n     Noise measurement is active.    \n     Turn Noise measurement off.     
 12 \n    Only dBm and dBV can be used   \n while Noise measurement is active.  
 13 \n    Frequency Counter is active.     \n     Turn Frequency Counter off.     
 14 \n     Delta marker is not active.     \n      Turn the Delta marker on.      
 15 \n     External mixer is selected.     \n     Set the mixer to Internal.      
 16 \n    Mixer band cannot be changed     \n   while in Separate-screen mode.    
 17 \n          Not available in           \n        Separate-screen mode.        \n      Set to Full-screen mode.       
 18 \n View/Blank Trace mode is selected.  \n  Set the Trace mode to Write mode.  
 19 \n      Trigger source incorrect.      \n      Set the trigger source to      \n            Video or IF.             
 20 \n    Marker not on selected trace.    \n       Move the marker first.        
 21 \n       Signal Ident is active.       \n       Turn Signal Ident off. 
 22 \n     Scale not set to 10dB/div.      \n      Change to 10dB/div first.      
 23 \n\n     Parameter is out of range.      
 25 \n\n Calculated power is off the scale.  
 26 \n          Editor is active.          \n       Quit the editor first.        
 27 \n\n  Frequency table contains no data.  
 28 \n       No Cal signal detected.       \n      Check the CAL OUT signal.      
 30 \n Not available for OBW measurements. \n           Quit OBW first.           
 31 \n     Not available for Harmonics     \n            measurements.            \n        Quit Harmonics first.        
 32 \n     Not available for Spurious      \n            measurements.            \n        Quit Spurious first.         
 33 \n Not available for ACP measurements. \n           Quit ACP first.           
 34 \n  Not available for ACP graph mode.  \n      Quit ACP graph mode first.
 35 \n  Eye Opening measurement is active. \n  Turn Eye Opening measurement off.     
 36 \n  Only available in Separate-screen  \n     mode using Sample Detector.     
 37 \n     Internal mixer is selected.     \n     Set the mixer to External.      
 38 \n     Loss vs. Freq is turned on.     \n      Turn the correction off.       
 39 \n     Trace average is operating.     \n       Turn Trace average off.       
 41 \n     Trace Point is set to 501.      \n           Change to 1001.           
 42 \n           Not available.            \n         Turn off Zoom mode.         
 43 \n\n           No trace data.            
 44 \n    Attenuator is set to manual.     \n          Select Auto mode.          
 45 \n The active marker is out of range.  \n         Move the marker or          \n    check the search conditions.     
 46 \n       Not peak points found.        \n    Check the search conditions.     
 47 \n\n   There are no more peak points.    
 48 \n    Trace Max/Min Hold is active.    \n       Turn Max/Mix Hold off.        
 49 \n    Normal marker is not active.     \n     Turn the Normal marker on.      
 52 \n         Currently sweeping.         \n  Stop the sweep before proceeding.  
 60 \n  Vertical scale is set to dB/div.   \n      Set the scale to Linear.       
 61 \n Not available for C/N measurements. \n           Quit C/N first.           
 62 \n    Not available for Phase Jitter   \n            measurements.            \n       Quit Phase Jitter first.        
 63 \n  Not available for IM measurements. \n         Quit IM Meas. first.           
 66 \n\n     Not available in F/T mode.      
 67 \n\n     Not available in F/F mode.      
 68 \n\n     Not available in T/T mode.      
 69 \n\n Not available in Delay sweep mode.  
 70 \n\n     Display line is not active.     
 71 \n\n     Limit Line 1 is not active.     
 72 \n\n     Limit Line 2 is not active.     
 73 \n\n     No limit lines are active.      
 74 \n         Invalid data mode.          \n        Set to Relative mode.        
 75 \n           Not available.            \n       Set to F/T or F/F mode.       
 76 \n      No 3rd order peak found.       \n     Check the search conditions.
 77 \n\n   This function is not available.   
 78 \n\n  This function is already active.   
 79 \n     Trace Normalize is active.      \n      Turn Trace Normalize off.      
 80 \n\n Not available in Gated sweep mode.  
 81 \n\n Not available in Manual sweep mode. 
 82 \n\n Not available in Window sweep mode. 
 83 \n       Not available in either       \n Manual sweep or Window sweep mode.  
 84 \n           Not available.            \n     Gate source is EXT Gate In.     
 85 \n           Incorrect data.           \nSet span to (1.0 + alpha)*Tf or more.
 86 \n\n       Incorrect parameter(s).       
 87 \n   Root Nyquist filter is active.    \n        Tunr the filter off.         
 88 \n   Separate-screen mode is active.   \n     Change to Full-screen mode.     
 90 \n           Not available.            \n        Set XY anchor first.         
 91 \n           Not available.            \n       Delta marker is active.
 95 \n           Not available.            \n  Turn off Artificial Analog mode.   
 96 \n           Not available.            \n   Turn on Artificial Analog mode.   
100 \n            Not available            \n       in High speed ADC mode.       
105 \n            Not available            \n      in Frequency Domain mode.      
110 \n            Not available            \n      in Continuous peak mode.       
111 \n            Not available            \n    in Continuous XdB Down mode.     
112 \n       Not peak signal found.
115 \n Trigger source is not VIDEO or EXT. \n    Set trigger to VIDEO or EXT.     
116 \n      Trigger source incorrect.      \n      Set the trigger source to      \n         EXT/IF/RF trigger.          
120 \n           Not available.            \n       Change the sweep time         \n        to less than 500sec.     
125 \n\n Not available in Digital RBW mode.      
126 \n      Reached the limit of span      \n        in Digital RBW mode.      
127 \n           Not available.            \n Change span to less than the limit.
300 \n        Printer is not ready.        \n        Check printer setup.         
301 \n       Printer cable problem.        \n        Check printer cable.         
302 \n\n       Printer not responding.       
303 \n          Plotter problem.           \n       Check power and cable.        
304 \n\n          SIO port is busy.          
305 \n\n       Input buffer overflow.        
306 \n\n     Not available in this mode.
400 \n\n        Input ATT Cal failed.        
401 \n\n       IF Step AMP Cal failed.       
402 \n\n      Log Linearity Cal failed.      
403 \n\n       Total Gain Cal failed.        
404 \n\n      RBW Switching Cal failed.      
405 \n\n      Amplitude MAG Cal failed.      
406 \n       Insufficient Cal data.        \n          Execute CAL ALL.           
407 \n\n         HS ADC Cal failed.          
409 \n\n       Normal ADC Cal failed.        
600 \n\n        Illegal parameter(s).        
601 \n\n    Illegal file or device name.     
602 \n   Incompatible firmware version.    \n    Data cannot be used with this    \n             instrument.             
603 \n\n        Cannot be formatted.         
604 \n\n      Cannot rename this file.       
605 \n\n         Corrupt file data.          
607 \n\n  Specified device does not exist.   
608 \n\n          No media present.          
609 \n\n        Directory not found.         
610 \n\n        File already exists.         
611 \n\n           File not found.           
612 \n\n    Invalid disk format (Type 1)     
613 \n\n           Read-only file.
614 \n\n            Disk is full.            
615 \n\n        Write-protected file.        
616 \n\n          Read-only media.           
618 \n\n   Invalid boot sector signature.    
619 \n\n             CRC error.              
621 \n Invalid Frequency-Correction data.  \n     Contact a service engineer.     
625 \n\n        Device name too long.        
626 \n\n         Extension too long.         
627 \n\n         Filename too long.          
628 \n\n         Pathname too long.          
631 \n\n             I/O error.              
633 \n\n    Invalid disk format (Type 2)     
634 \n\n Selected file or register is empty. 
635 \n   Communication system unmatched.
800 \n   The last process is in progress.    
##
## (( Modulation Option Message. 1999/2/5 )) ##
700 \n           System Error.             \n     Cannot allocate                 \n        the required memory.
701 \n           System Error.             \n     Clock is not operational.
702 \n   Modulation Gain CAL error.        \n     Check 30 MHz CAL signal         \n           for connection.
703 \n   Modulation DC CAL error.          \n  Remove input signals and try again.
704 \n           Time Out!                 \n        No Trigger Detected.
705 \n   Input Level is out of Range.      \n       Check the Ref. level.
706 \n           No graph data.            \n       Execute measurement.
707 \n       Input level is too low.       \n       Adjust the Ref. level.
708 \n           System Error.             \n     Contact qualified engineer.
709 \n    Incorrect channel settings.      \n  Reset the channel class and number.
710 \n      Auto Level completed !
711 \n  Auto Level Set can not be succeed. \n     Signal level is not stable.
712 \n    Cannot execute measurement.      \n       Because  is too low.
713 \n   Cannot synchronize to Pilot.      \n          Adjust PN Offset.
714 \n    Cannot converge equation.        \n          Adjust threshold.
715 \n       Frequency Error is            \n         out of Meas. Range.
716 \n    Cannot execute measurement       \n     in the selecetd Meas Mode.      \n         Change Meas Mode.
717 \n    Cannot execute measurement.      \n    Set Sync Type to                 \n           SYNC WORD (UNIQUE WORD).
718 \n      Sync Word (Unique Word)        \n         is not detected.            \n   Check Sync Word (Unique Word).
719 \n  Burst signal is not detected.      \n  Check Burst length or Ref. level.
720 \n      Data detection error.          \n      Check the input signal.
#
721 \n  Modulation Gain CAL error! (#100)  \n     Check 30 MHz CAL signal         \n           for connection.
722 \n  Modulation Gain CAL error! (#200)  \n     Check 30 MHz CAL signal         \n           for connection.
723 \n  Modulation Gain CAL error! (#300)  \n     Check 30 MHz CAL signal         \n           for connection.
724 \n  Modulation Gain CAL error! (#110)  \n     Check 30 MHz CAL signal         \n           for connection.
725 \n  Modulation Gain CAL error! (#120)  \n     Check 30 MHz CAL signal         \n           for connection.
726 \n  Modulation Gain CAL error! (#210)  \n     Check 30 MHz CAL signal         \n           for connection.
727 \n  Modulation Gain CAL error! (#220)  \n     Check 30 MHz CAL signal         \n           for connection.
728 \n  Modulation Gain CAL error! (#310)  \n     Check 30 MHz CAL signal         \n           for connection.
729 \n  Modulation Gain CAL error! (#320)  \n     Check 30 MHz CAL signal         \n           for connection.
#
730 \n          Cannot measure             \n    Multi-Burst/Continuous Signal.   \n       Set Meas Mode to BURST.
731 \n      Cannot detect Sync Word.       \n   Check link or syncword number.
732 \n     Sync Word position              \n          is different from STD.
733 \n       Input Level is too Low.       \n       Adjust Ref. level,            \n        trigger delay, burst type.
734 \n            Result Error.            \n   Check input signal or settings.
735 \n   Cannot measure Baseband signal.   \n          Set Input to RF.
736 \n     Cannot measure GMSK signal.     \n       Set Modulation Type to        \n          3PI/8 Shift 8PSK.
737 \n   Cannot measure 3PI/8 Shift 8PSK   \n   signal. Set Modulation Type to    \n                GMSK.
#
738 \n    Cannot execute measurement.      \n   Because Meas Unit is too small.
739 \n    Incorrect channel settings.      \n  Reset the channel SF and number.
#
740 \n   Cannot measure baseband signal.   \n   This function is available to     \n            RF input only.
741 \n      Burst length is too short      \n    to measure with STD sequence.    \n    Fill the payload with the data.
742 \n    This function is available to    \n           DOWNLINK only.            \n         Set Link to DOWNLINK.
743 \n     Cannot allocate sufficient      \n               memory.               \n     Set Power Unit to RELATIVE.
744 \n    No Idle Slot within a frame.     \n       Check the input signal.
745 \n   No Active Slot within a frame.    \n       Check the input signal.
746 \n   Cannot find out active Channel.   \n       Down the MAC Threshold.
747 \n   Setup does not suit the signal.   \n          Check the setup.
#
750 \n  Handshake error occurred to DSP.   \n     Contact qualified engineer.
751 \n    Cannot detect Mod. DSP board.    \n     Contact qualified engineer.
752 \n Invalid Level Cal Correction data.  \n     Contact qualified engineer.     
753 \n  Check Sum error occurred to DSP.   \n     Contact qualified engineer.
#
#=== HDR ===
760 \n  Level of MAC channel is too low.   \n       Check MAC Channel.
#
##=== cdma200 ===
780 \n  Cannot execute measurement.        \n  Check the QOF and Data Rate        \n          of the Channel Def.Table.
781 \n  Incorrect Channel Def. Table       \n                         settings.   \n  Check the Channel Def. Table.
782 \n  Cannot synchronize to PICH.        \n         Adjust Threshold.
783 \n  Cannot synchronize to PICH.        \n         Adjust PN Delay.
#
795 \n           System Error.             \n      Memory test failed. (#0)
796 \n           System Error.             \n      Memory test failed. (#1)
797 \n           System Error.             \n      Memory test failed. (#2)
798 \n           System Error.             \n      Memory test failed. (#3)
##
##((( for TG Option )) ########################
310 \n        No TG signal detected.       \n        Check TG OUTPUT signal.
311 \n          Level Cal failed.          \n      Check connection between       \n         TG OUTPUT and INPUT.
312 \n      TG Power Sweep Cal failed.     \n      Check connection between       \n         TG OUTPUT and INPUT.
313 \n      Through Correction failed.     \n      Check connection between       \n         TG OUTPUT and INPUT.
314 \n       Invalid Level Cal data.       \n        Execute TG Level Cal.
315 \n    Invalid Power Sweep Cal data.    \n     Execute TG Power Sweep Cal.
316 \n   Invalid Through Correction data.  \n     Execute Through Correction.
317 \n      TG frequency out of range.
320 \n        Px dB point not found.
810 \n      Preselector set to 3.6GHz.
##
##
##((( for Rx Test Control[R3560/61/62] )) #####
650 \n\n        Rx Tester timed out.
651 \n\n            Sync error.
652 \n\n            Clock error.
653 \n\n      SensPoint is not found.
656 \n\n    CDMA test source timed out.
#660 \n\n          R3561 timed out.
661 \n      Invalid connected unit.        \n      Please connect other unit.
662 \n\n   Not available in this revision.
663 \n\n         Self Test failed.
664 \n\n        Calibration failed.
805 \n\n        Self Test completed.
806 \n\n       Calibration completed.
807 \n\n       Calibration timed out.
808 \n   Since the system was switched,    \n         R3562 is re-started.